Shop the Holiday Gift Guide
Medical Equipment
Neither iFixit nor this repair information is endorsed by or affiliated with the medical device manufacturer.
Documents
Background
The Respironics BiPAP Focus (Model Number:8-400258-00), is a non-invasive ventilatory system that was released in 2006. The Respironics BiPAP Focus is a hospital-grade ventilator that includes features such as a large screen, backup batteries, adjustable alarms, and nurse calls. This device is distributed worldwide and is meant for hospital use, but can also be used in home care.
In 2008 on September 22nd, the manufacturer issued a Class 1 recall on the device, citing a power supply failure related to power cord connection and possible power surges. The recall status was terminated on July 10th, 2010.
Identification
This ventilator has a large screen interface. It is constructed with a white/grey plastic exterior and weighs in at 10 lbs (4.5 kg). The device is portable utilizing a smaller design with a width of 11.4 inches, a depth of 14 inches, and a height of 5.5 inches. The manufacturer and model name is printed on the front of the device near the bottom of the screen.
Source: Class 1 Recall Status
Technical Specifications
Controls
  • CPAP: 4-20 cm H2O
  • EPAP: 4-25 cm H2O
  • IPAP: 4-30 cm H2O
  • Rise Time: 1-6 (1 = 0.1 s and 6 = 0.6 s)
  • Rate: 1-30 BPM (accuracy: ±BPM or ±10% of setting)
  • I-time: 0.5-0.3 s
  • Ramp Time: 0-45 min
  • Ramp Start: 4 cm H20 to EPAP
  • Maximum Flow Capacity: 160 l/min
  • Leak Compensation: Up to 60 l/min
Dimensions
  • Height: 5.5” (14 cm)
  • Width: 11.4” (29 cm)
  • Depth: 14” (35.5 cm)
  • Max Weight: 10 lbs (4.5 kg)
Patient Parameters
  • Patient Circuit Pressure: 0-35 cm H2O
  • Rate Range: 0-99 BPM
  • Estimated VT Range: 0-4000 ml
  • Estimated MV Range: 0-99 l/min
  • Patient Leak Range: 0-200 l/min
  • Apnea Rate Range: 0-99 per hour
Alarms
  • High System Pressure: 5 cm H2O over target IPAP pressure
  • Low System Pressure: 5 cm H2O under target EPAP pressure
  • Apnea Range: 20, 40, 60s or off
  • Apnea Rate Range: 5, 10, 20 apneas per hour or off
  • Nurse Call: Normally open, normally closed
Electrical
  • Input Rating: 100 VAC / 230 VAC
  • Battery Backup: 3.8 amp-hour (NiMH)
  • Fuse: No replaceable fuse
  • Nurse Call: 0.280 A (rated current); 250 V (rated voltage)
Source: Respironics BiPAP Focus Specs
Additional Information
FDA Product Recall Status Document
Soma Technologies Product Specifications
Philips Respironics BiPAP Focus Product Page
BiPAP Focus Brochure
BiPAP Focus Ventilator Service Manual
BiPAPフォーカス 医療機器添付文書(第3版)
Author
with
iRobot
1 Reputation
632 Guides authored
Past 24 Hours: 2
Past 7 Days: 13
Past 30 Days: 52
All Time: 1,276
